Michael: You can calibrate a load cell in a similar fashion: Build a
rigid lever arm of known length. Mount it to the load cell such that it
sticks out in a perpindicular fashion from the load cell axis, parallel
to the ground. Zero out your gages to account for the added weight of
the lever arm. Apply a load of known force, (20 lbs or so,) and
calibrate your sensor for torque. If we assume your lever arm is three
feet long, and you have placed a 20 pound weight on it, you now have 60
ft-lbs (Lbs-ft?) of torque on your cell. Again, assume that the gages
are linear to account for higher amounts of torque.
If 5V = 1200 ft.lbs, and zero ft-lbs equals 2.5V, you have 2mv per
ft-lb, or a cal value of 2.5V plus 120mv, or 2.620mv predicted output of
your load cell gaging.
Cheers,
Dan Burk
Instrumentation Engineer for oh, about 12 months now.